Removal procedure
1. Turn the fastening ring on each sensor head connection cable connector in the
direction of the arrow to loosen.
1
2. Grasp each connector on the sensor head connection cable and pull out to re-
move.
2
When disconnecting, always make sure that the fastening ring has been complete-
ly loosened before pulling out the cable. Risk of damage if you pull the cable with
excessive force (15N or more) with the fastening ring tightened.
6
BEAM ALIGNMENT AND REFERENCE WAVEFORM REGISTRATION
● When installing the product, always register a reference waveform. If a reference
waveform is not registered, correct measurement will not be possible.
● Register the reference waveform without a measurement object.
● When detecting a measurement object with a regular reflection component such
as glass or a mirror surface, reflected light from the measurement object may
prevent correct detection. In this event, mount the emitter and receiver pair at an
angle so that the reflected light does not enter the emitter or reflector. After ad-
justing the angle, always perform beam alignment.
<Beam alignment>
● Check the digital display of the controller as you align the beam.
<Reference waveform registration>
● You can perform reference waveform registration by pressing the ENTER key on
the controller with the beam correctly aligned. The reference waveform is stored
in EEPROM in the receiver.
7
CAUTIONS
The HG-T□ Sensor Head is designed to be used in combination with an HG-TC□
Controller. If used with a separately purchased dedicated controller, the specifica-
tions will not be met and product malfunctioning or damage may occur.
● This product has been developed / manufactured for industrial use only.
● The receiver in this product uses an EEPROM. The EEPROM has a service life of
one million setting operations.
● Do not use this product outside the range of the specifications. Risk of an accident
and product damage. There is also a risk of a noticeable reduction of service life.
● To ensure performance, do not use the product after turning ON the power until it
has warmed up for at least 30 minutes.
● Do not wire in parallel with a high-voltage line or power line, or run through the
same conduit. Risk of malfunctioning due to induction.
● Do not apply stress such as excessive bending or pulling to the extracted part of a
cable.
● Do not touch or allow any foreign objects to enter the terminal inside the connector.
● Do not allow ambient light such as sunlight to directly enter the sensor head re-
ceiver. When precision is particularly required, take measures such as attaching a
shield to the sensor head.
● Avoid use in a location subject to sudden temperature changes that cause con-
densation.
● This product is suitable for indoor use only.
● Avoid dust, dirt, and steam.
● Do not use in a location where there are corrosive or other harmful gases.
● Ensure that the product does not come into contact with organic solvents such as
thinner.
● Ensure that the product does not come into contact with strong acid or alkaline.
● Ensure that the product does not come into contact with oil or grease.
● The product cannot be used in an environment that contains flammable or explo-
sive gases.
● Performance may not be satisfactory in a strong electromagnetic field.
● Do not use the product in a location subject to intense vibration or shock.
● The connectors are not structurally dustproof, waterproof, or corrosion-resistant.
Do not use the product underwater or in the rain. Take care of the usage environment.
● This product is a precision device. Do not drop or otherwise subject to shock. Risk
of product damage.
● Never attempt to disassemble, repair, or modify the product.
● When the product becomes unusable or unneeded, dispose of the product appro-
priately as industrial waste.
1
2
2
2
8
SPECIFICATIONS
Type
Model No.
Combined controller
Position detection method
Installation distance
Measurement width
Red semiconductor laser Class 1 [JIS / IEC / GB / FDA (Note 1)]
Light source
Maximum output: 0.3mW, Peak emission wavelength: 655nm
Repeatability (Note 2)
Linearity (Note 3)
Minimum sensing object (Note 4)
Temperature characteristic
(Note 5)
Protection
Pollution degree
Ambient temperature
-10 to +45°C (No condensation or icing), Storage: -20 to +60°C
Ambient humidity
Ambient illuminance
Incandescent light: 5,000lx at the light-receiving face (Note 6)
Operating altitude
Grounding method
Material
Case: Aluminum die-casting, Light emitting / receiving faces: Glass
Cable
Attached 0.2m shielded cable with 0.08mm
Weight
Specification figures are based on the digital measurement values obtained with a sensor combined with an HG-TC□
controller.They were measured under the following condition unless otherwise specified.
• Ambient temperature : +20°C
• Average controller count setting: 16 times
• Measurement object : Opaque knife edge
• Measurement object position condition: Light half-blocked at the center of 100-mm installation distance
Notes: 1) This is based on the FDA Standard, according to Laser Notice No. 50 of the FDA Standard.
2) P-P value of deviations in digital measurements obtained with light half-blocked at the center of installation
distance
3) Indicates the error from the ideal line formed by digital measurements.
4) When light is light blocked at the center of 500-mm installation distance.
5) When light is light half-blocked at the center of 100-mm installation distance.
6) When the sampling cycle of the controller is set to "Standard Sampling". For detailed information, refer to the
"HG-T Series User's Manual" (our Website: https://panasonic.net/id/pidsx/global ).
7) Do not use or store in an environment that has been pressurized to an air pressure higher than the atmo-
spheric pressure at 0m.
Panasonic Industry Co., Ltd.
Panasonic Industrial Devices SUNX Co., Ltd.
https://panasonic.net/id/pidsx/global
Please visit our website for inquiries and about our sales network.
Panasonic Industrial Devices SUNX Co., Ltd. 2022
September, 2022
Measurement width 10 mm / slim type
HG-T1110
HG-TC□
CMOS method
0 to 500mm
10mm
1µm (at an installation distance of 20mm)
2.5µm (at an installation distance of 100mm)
5µm (at an installation distance of 500mm)
±0.12% F.S. (at an installation distance of 20mm)
±0.28% F.S. (at an installation distance of 100mm)
ø0.5mm (at an installation distance of 500mm)
±0.03% F.S./°C
IP67 (IEC) (excluding the connector)
2
35 to 85% RH, Storage: 35 to 85% RH
2,000m or less (Note 7)
Capacitor grounding
2
4-conductor type, round connector
Emitter : 30 g approx., Receiver : 25 g approx.
PRINTED IN JAPAN