Frequency measurement and switching instruments, single channel tachometer with display and 0/4-20ma output, +14v/5v sensor supply, single channel tachometer with display and 0/2-10v output, +14v/5v sensor supply
Operating Instructions T411 / T412 JAQUET AG Contents: SAFETY INSTRUCTIONS PRODUCT FEATURES SPECIFICATIONS General Inputs 3.2.1 Analog Sensor connection (Sign) 3.2.2 Digital Sensor connection (IQ) 3.2.3 Binary input Outputs 3.3.1 Analog output 3.3.2 Relay 3.3.3 Open Collector Output Data communication 3.4.1...
Page 3
Operating Instructions T411 / T412 JAQUET AG OPERATING BEHAVIOR Power on 9.1.1 Analogue Output 9.1.2 Relay Output Measurement 9.2.1 The adaptive Trigger level 9.2.2 Signal failure Functions 9.3.1 Limits and Window Function 9.3.2 Parameter set A and B 9.3.3 Relay hold function 9.3.4...
Series T400 tachometers measure and monitor frequencies (speed proportional values) in the range 0 to 35,000 Hz. The following features are available: • 1 Current or voltage output (T411 - current, T412 - voltage) • 1 Sensor frequency output •...
Operating Instructions T411 / T412 JAQUET AG 3.2 Inputs 3.2.1 Analog Sensor connection (Sign) Frequency range (-3dB) 0.01 Hz / 35 kHz Input impedance 30 kOhm • Input voltage Max. 80V Max. frequency against input voltage O.K. NOT O.K. Trigger: 500mVpp Trigger: 20mVeff 0.01...
Operating Instructions T411 / T412 JAQUET AG Sensor monitoring 1 of 3 settings may be configured via software: • No Sensor Monitoring • Monitoring of powered sensors [Also for 2 wire sensors supplied via the Pull-up resistor (Jumper J1) ].
Operating Instructions T411 / T412 JAQUET AG 3.3 Outputs 3.3.1 Analog output T411 T412 Type Current Voltage 0…20 / 4…20 mA 0…10 / 2…10 V Load Max. 500 Ohm Min. 7 kOhm, Max. 1.4mA Open circuit voltage Max. 13V Operating mode 10.5...
Operating Instructions T411 / T412 JAQUET AG 3.3.2 Relay Type Mono-stable change-over Programmable – 1 lower and 1 upper set point per limit. Limit Hysteresis Functions 2 programmable parameter sets selectable via binary input • Reaction to Alarm, Sensor fault, Limit, always on or off.
Operating Instructions T411 / T412 JAQUET AG 3.5 Environment 3.5.1 Climatic conditions Standard KUE in accordance with DIN 40 040 Operating temperature - 20 ... + 70 °C Storage temperature - 20 ... + 70 °C Relative humidity 75% averaged over the year; up to 90% for max 30 days.
Operating Instructions T411 / T412 JAQUET AG 4 Principle of operation 4.1 General T400 tachometers are controlled by a microprocessor. They work according to the period measurement principle whereby the input period is measured with subsequent computing of the reciprocal value corresponding to the frequency or speed.
Operating Instructions T411 / T412 JAQUET AG 4.2 Machine factor The machine factor establishes the relationship between sensor frequency and corresponding speed. Machine factor Signal frequency at machine speed n Machine speed There are 2 ways of determining the value: 4.2.1 Known (Measured)
CEC and the Canadian deviations in the standard, including overcurrent and fault protection as required. 6 Connnections 6.1 Front view The T411 / T412 display along with the RS232 interface and the status LED are located at the front. Communications via RS232 are described in section 8.2. Revision: 003...
Operating Instructions T411 / T412 JAQUET AG 7 Hardware configuration 7.1 Analog Sensor input (Sign) Jumper position J1: Sensor type J2: Adaptive trigger level range 2 wire sensors 28mV to 6.5V (>20mV (with 820Ohm Pull Up resistance) 3 wire and electromagnetic sensors 250mV to 6.5V...
The tachometer RXD must be connected to the PC’s TXD and vice versa. T411 / T412’s do not use a standard RS232 signal (-5V…+5V) but operate at 5V CMOS levels, compatible with most PC’s as long as the cable is not longer than 2m.
Operating Instructions T411 / T412 JAQUET AG 8.4 Parameter list and ranges If you already have a configuration file you can open and view it using the T400 Windows Software menu File Open You can also connect the T400 to a PC (see section 8.2) and read back the parameters, T400 ...
Operating Instructions T411 / T412 JAQUET AG 8.5 Parameters Parameters are changed in the sub menus from the drop down menu „Configuration“. Warning: New configurations only become active after being uploaded into the T400 via: T400 Write Parameters 8.5.1 System parameters (Configuration System) Machine factor The machine factor establishes the relationship between sensor frequency and associated speed.
In the case of a negative transfer function the end value must be set smaller than the start value. Output range 0…20mA or 4…20mA for the T411. 0…10V or 2…10V for the T412. Output time constant The analog output signal may be smoothed by applying a software time constant. This damping is deactivated when the time constant is 0.0 seconds.
Operating Instructions T411 / T412 JAQUET AG 8.5.4 Limit (Configuration Limit) The T400 series offers 2 independent limits Limit 1 and 2. Status Limits are selected here. If the limit is deactivated, the other values such as set points and mode have no further effect.
Operating Instructions T411 / T412 JAQUET AG 9 Operating behavior 9.1 Power on 9.1.1 Analogue Output After switching-on the tachometer, the output relates the lower value of the defined output range. After completion of the first measurement the output goes to the corresponding measured value.
Operating Instructions T411 / T412 JAQUET AG 9.2.2 Signal failure In the event of a sudden loss of a good signal, no positive edge arrives to complete the measurement or start a new one. Once the minimum measurement time (Fixtime) has lapsed, the unit waits for twice the last measurement period, following which half the last measured speed is assumed.
Operating Instructions T411 / T412 JAQUET AG 9.3.4 Binary Input 2 functions are executable using the binary input: • Switching between parameter sets A and B. See 9.3.2 Parameter set A and B. • Resetting a latched relay. See 9.3.3 Relay hold function.
Operating Instructions T411 / T412 JAQUET AG 9.6 Display adjustments 9.6.1 Brightness The brightness of the display can be adjusted by pressing the up and down arrow buttons at the front panel. 9.6.2 Contrast The contrast of the display can be adjusted by pressing the up or down arrow button and the center button at the front panel.
For software display version 1.2 are values to 999.9k possible. 14 Warranty The standard warranty in the event of a manufacturing failure confirmed by Jaquet consists of repair or replacement within 12 months of delivery. Ancillary costs are excluded s well as damages caused by use outside of the specification.