FRU Replacement
Leakage Tests
GE Healthcare recommends that you perform the electrical leakage tests whenever
internal assemblies are serviced.
See
"Performing Leakage Tests" on page
Shroud Removal
To remove the T2100 Treadmill shroud for the internal visual inspection or for service:
• Disconnect the power cord from the rear connector panel.
• Remove the screws on the shroud back and remove the back panel.
• Remove the 2 bolts on each side of the shroud.
• Lift off the shroud.
Item
1
2
106
Description
Back Panel
Shroud Side Bolts
T2100 Treadmill
47.
2021403-120F