8. Set the temperature bath to the maximum range of the TT820D transmitter
( Ex: 200
o
C = 20 mA )
9. With the temperature standard, verify the bath temperature and enter the
value in the " Second Point " text box located in the " 2 Points Calibration "
section of the " Programming " window.
10. Immerse the TT820D probe in the temperature bath and wait for the
output stabilizes.
11. Click on the " Read " button associated to the " Second Point " calibration.
12. Click on the " Process " button, located in the " 2 Points Calibration "
section of the "Programming" window, to update the calibration
parameters in the TT820D.

7. Calibration with a thermocouple simulator

Both calibrations, one point calibration and advanced calibration, can be
performed with a thermocouple simulator. In both cases, proper thermocouple
wires should be used to connect the simulator to the TT820D transmitter.
For the one point calibration, simply replace the probe ( thermocouple ) by the
simulator and set the output of the simulator to mid-range of operation. Wait
for output signal stabilizes and adjust, with the " UP " and " DW " push buttons,
the output to 12.00 mA.
For the two points calibration, simply replace the probe ( thermocouple ) by the
simulator and follow the same procedure as for the " Advanced Calibration ".
Instead of using a temperature bath, adjust the output of the simulator to the
minimum and maximum values of the range.

8. Specifications

Power Supply :
Input :
Output :
Supply Voltage Effect :
Load Capability :
Zero Drift :
Span Drift :
Ambient Operating Temperature :
Warm-up time :
Accuracy :
K, E, J, T :
Cold Junction Compensation Error :
K :
J :
T, E :
4
12-36V DC, polarity protected
Thermocouples type J, K, T, E
4-20 mA
± 0.02 % / V
Rmax.= ( Vsupply - 12V ) / 20 mA
± 0.1 % / °C
± 0.1 % / °C
- 40 to +80 °C ( -40 to +158 °F )
30 sec.
± 0.5 % of FS
± 1 °C max. at - 20 to 50 °C
± 2 °C max. at - 20 to 50 °C
± 3 °C max. at - 20 to 50 °C
2511, Guenette Montreal ( Quebec ) Canada H4R 2E9
Tel.: [ 514 ] 337-7471
www.intempco.com
INTEMPCO's TT820D
Digital Thermocouple Transmitter
INSTRUCTION MANUAL

1. Description

The TT820D is a digital transmitter that converts the signal from a thermocouple
to a linearized 4-20 mA output. The TT820D features a one point calibration
adjustment with two push buttons as well as an advanced software programming
via the RS-232 input. The advanced programming allows a two points calibration,
a re-scaling, filtering options and identification of the transmitter.

2. Note

Prior to unpacking and installation, please read the operating instructions and
follow them carefully. These units are to be used, serviced, and repaired only by
individuals who are familiar with the operating instructions and the applicable
regulations for operational safety and accident prevention.

3. Control of Units

The units are calibrated and checked before shipment and shipped in good
conditions. If you detect a visible defect on the unit, we recommend that you
carefully check the packing material. In the event of a defect, please immediately
notify the mail service / freight forwarder, as they are responsible for shipping
damage.
Fax : [ 514 ] 337-7475
1

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the TT820D and is the answer not in the manual?

Questions and answers

Summary of Contents for Intempco TT820D

  • Page 1: Specifications

    2511, Guenette Montreal ( Quebec ) Canada H4R 2E9 8. Set the temperature bath to the maximum range of the TT820D transmitter Tel.: [ 514 ] 337-7471 Fax : [ 514 ] 337-7475 ( Ex: 200 C = 20 mA ) www.intempco.com...
  • Page 2: Electrical Connections

    The connections should be made in the following order : Interface 1. Connect the Thermocouple to the « C » and « D » terminals on the TT820D. Be sure to respect the polarity of the thermocouple ( C to Neg. D To Pos. ) 1.

Table of Contents